Can cable movement cause the distance counter to become inaccurate?
Picked up a used push rod reel with a mechanical counter on the frame, numbers turn as the cable feeds through. I've been using it to log lateral lengths on a few jobs and the numbers don't match what I measure after the fact. Not by a little either, sometimes 8-10 ft off over a 150 ft run. Seems worse when I'm pulling the cable back out than when I'm pushing in. Cable is not kinked, reel sits level. Is the counter just junk, or is there something about how the cable moves through the roller that throws it off?
Two rollers or one? If it's a single wheel pressed against the cable, the cable has to actually turn the wheel. If it rides up on the flange or slips off to the side for a stretch, it counts nothing for that stretch and you come up short.
Fast retrieval does it every time. Wheel can't keep up with the cable and the cable just Slides past it.
On my old reel the counter runs off a wheel with a rubber tire on it. Once that tire glazes over or picks up grease, it slips and you lose footage. Cleaned it with alcohol and the readings came back close to right. Also had a stretch where the wheel was fine but the cable was wet and slick coming out of a flooded line, same result. Now I wrap tape on the cable at the start and count from there if I need a real number.
Counters are a rough guide. If the number is going in a report, I measure it. Tape on the rod at intervals is cheaper than being wrong.
Which reel is it? Some of the cheaper frames have the counter running off a plastic wheel that wears a flat spot within a season.
Does your counter read off a wheel the cable touches, or off the drum itself? I've seen reels where the counter is driven from the drum, and any change in how the cable wraps changes the number.
There are two different errors here and people mix them up. Slippage between the cable and the counter wheel is random, so it shows up as small losses that don't scale with the length of the run. A worn or undersized counter wheel is systematic: the circumference is smaller than nominal, so every revolution is short and the error grows with distance. If your error grows with the length of the run, look at the wheel. If it's erratic and only happens on some passes, look at slippage.
Can the cable itself stretch? Older cable under a lot of pull has to have some give in it over time.
Single wheel pressed against the cable with a spring arm, runs on top of the cable as it comes off the drum. Runs are 100-200 ft mostly, 3 and 4 inch lines, sometimes 6. I'll do the tape mark tomorrow and compare the two numbers. Cleaning the wheel first, it does look glazed.
Spring arm tension matters more than people think. If the spring is tired the wheel only loses contact when you pull hard, so pushing reads close and pulling reads short. That matches what you described.
Anybody accounted for the cable taking a helix in the pipee? Pushing through a bend or into a bigger line, the cable doesn't lie straight. The footage along the cable is longer than the pipe, so the counter overstates how far down the line the head is. On a 100 ft run with a couple of bends it adds up.
The coil thing is real. You can watch the head stop movin while the counter keeps climbing.
Counter accuracy is not the same thing as location accuracy. Even a perfect counter tells you cable lentgh, not distance down the pipe.
One percent on 200 ft is two feet. Two feet is the difference between the tap and the main. That's why I gave up on the number and mark the rod.
How do you calibrate one of these? Pull a known length past the wheel and see what it reads?
Pull ten feet, note the reading, repeat three or four times. But I still don't buy wheel wear explaining 8-10 ft in 150. That's a big loss unless the wheel is badly chewed or the cable is skipping a lot of the time. Did yoou confirm the cable stys in the groove for the full pass?
Post a reply
Trending Discussions
- What is the best way to approach a tight 90-degree turn? 15 replies
- Are Vivax camera heads compatible with other inspection systems? 13 replies
- Can I connect a generic monitor to most sewer camera control units? 8 replies
- Do you mark the pipe route as you push the camera forward? 16 replies
- I started offering a camera check as part of my drain cleaning jobs 12 replies